    1. Before leaving the chamber the incubator doors and the inner Air Lock door must be
    closed.
    2. Depress the GAS foot pedal to allow the sleeves to fill with gas. This procedure
    keeps the sleeves from collapsing as you move outward, making it easier to
    reposition the doors.
    3. Hold the locking bars in a horizontal position, and close the doors. Rotate the locking
    bars to a vertical position, and gently tighten both armport door knobs, just until they
    “grab”.
    4. Before removing arms from the sleeves:
    A) Check to be sure that armport doors are securely sealed by slowly pushing
    both arms forward (this action generates a pressure within the sleeves).
    B) If the Manometer bubbles when the arms are pushed forward, this indicates
    that the doors are improperly sealed. Loosen the armport door knobs again,
    remove the doors. Check the O-rings for any particulate matter on the sealing
    areas, and repeat Step “A”. If Manometer remains stable, the doors are
    sealed properly.
    5. When the armport doors are properly sealed, slowly and carefully withdraw both
    arms from the sleeves.

 