Bactron User Group > Incubator Temperature, Safety, and Callibration » How do I calibrate the incubator?
1 to 1 of 1
-
- CommentAuthorShelly (Sheldon MFG)
- CommentTimeJan 5th 2009 edited
Compare the reading of the digital temperature display to an accurate reference
thermometer. If there is a difference, put the display into calibrate mode by pressing
both the Up and the Down buttons at the same time and holding them in for about
five (5) seconds or until the two outside decimal points start to flash on and off.
When the decimal points are flashing, the display can be calibrated to match the
reference thermometer by pressing the Up or Down button until the display reads
correctly.